manual calibration problem
nico
Posted: 15.10.2021 20:47
Modified: 07.10.2024 12:59
hello
impossible to manually calbrate aranet4.
The app crash here (see picture). When I select "start calibration", nothing happens...
Tanks for your help
aranet-techsupport
Posted: 18.10.2021 09:25
Hello Nico,
Please make sure that the calibration switch in Aranet4 battery compartment is set to the "Manual" position. Switch it to "AUTO" and then back to "Manual".
Repeat the calibration. if you still get the error, please contact support@aranet.com