2 sensors- differing readings
keithc
Posted: 19.05.2024 22:59
Modified: 21.11.2024 08:02
Hi, I have 2 aranet4's and I updated the firmware on each one. As they sit next to each other, their readings are vastly different- 1036 vs 652.
Should I do a manual calibration of one? Or both? Or, reset some other way? They each are factory calibrated so I'd hate to lose that if it's not necessary. Thanks.
aranet-techsupport
Posted: 20.05.2024 08:12
Hi!
Make sure you haven't accidentaly breathed on them and let the Co2 level settle, over an hour or so.
If the measurments are still more than ~60 ppm apart then do a manual calibration on both of them.